pH itself is not the only thing to consider concerning column lifetime. Which buffer do you use (organic high pH buffers are usually more column-friendly than inorganic; phosphate is supposed to be especially corrosive)? As already mentioned, which temperature? Lower temperatures will extend column lifetimes. Gradient or isocratic? Isocratic is more column friendly. Which ion-strength of buffer? Try to keep it low.
Be also aware that addition of organics will shift the pH of the buffer. (Yes, I know, pH is only defined in purely aqueous media...please don't start a thread on this again...)
I have made some pretty good experiences with the Waters XBridge and high pH-buffers. Stable like charm, even with 20mM phosphate pH11 at 50°C.
Disclaimer: I don't get any money from waters (unfortunately

).