In early 2010 I developed and validated an analysis for degradation products of gabapentin in an oral liquid pharmaceutical product using LC-ELSD.
The method was developed and validated in a very short timeframe as part of a licence submission.
The method was capable of detecting down to a 0.05% level of the stated dose and as such was at the limits of sensitivity forthe deetector.
As such it was never pretty but largely fit for purpose. After a year or two of running it the ELSD started to lose sensitivity, numerous service engineers visits later and the situation never improved (at the time it was the only ELSD within the company so comparison to other detectors was not possible).
Finally this year the company the company bought another detector.
Having got it set up I've immediately tried the method on it..................
At the moment even after optimisation of gas flows evaporator temp at best it gives half the response of the old detector ( originally we'd get a peak height of 45-50mV for the calibration Std on the new detector its 22mV at best).
So does anyone have any ideas as to whats happened?
Conidtions are as below
Agilent 1200 with Varian (now Agilent but was Polymer Labs 385 ELSD).
mobile phase 99:1 10mM Ammonium formate @ pH 3.2 with formic acid:Acetonitrile
Flow of 1ml/min with column at 50°C
Column is a 150x2.0x4 Synergi Polar RP
Thanks for any replies