I was just wondering if anyone has had trouble with high air and water peaks with the Varian 2000 GC/MS? Would anyone know the source of the problem? What was your plan of attack in correcting this? Any info would be great. Thank you.
-
- Posts: 114
- Joined: Mon Jul 18, 2011 6:10 pm
Check the injector end of the column, the injector oring, gas filter and the transferline end of column if you have small/medium leak, change the septum too.
If you have a large leak you would have already broken your filament i guess, and probably your turbo isn't even at full speed.
Davide
If you have a large leak you would have already broken your filament i guess, and probably your turbo isn't even at full speed.
Davide
Davide Balbo from Italy
2 posts
Page 1 of 1
Who is online
In total there are 63 users online :: 1 registered, 0 hidden and 62 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 62 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 62 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science