How do you calibrate a flow rate under the following conditions:
1 uL / min
calibration gives the RF
20 minutes or less
4-5 place accuracy
fully automated
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
lol I wish I was getting paid for this. I also forgot to mention that the process needs to be fully automated. I'll fix the OP to reflect this.Weigh a 2ml septum-capped vial to six figures. Insert the 1/16 inch outlet of the flow system through the septum and run for 20 min. Remove vial from tube. Reweigh. Calculate gain in weight. Correct for density. Repeat N times and take the mean. Send me 10% of your fee.
Peter
If you go this way, use plastic, low volume vials (that weigh less than 2g) if you want to use a semi-micro balance (many of which will not read beyond 2g, at least not to thousandths of a mg).Weigh a 2ml septum-capped vial to six figures. Insert the 1/16 inch outlet of the flow system through the septum and run for 20 min. Remove vial from tube. Reweigh. Calculate gain in weight. Correct for density. Repeat N times and take the mean. Send me 10% of your fee.
Peter