- Posts: 7
- Joined: Mon Jun 18, 2012 10:54 pm
I have been experiencing irregular shifting in my methods for a long time. I have recently replaced the pump seals, AIV catridge, ball check valve, purge valve frit filter, inline-frit filter, guard column, and capillaries for my HPLC. When running the tests, everything is fine, and the standard run was fine. But when I start running my sample in MeOH + 0.1%TFA and Water + 0.1%TFA gradient method, I get shifting, mostly elution time shifted later. I've decided to run the standard for this column again, and noticed that there was shifting in the forward direction compared to my 1st standard run when I just replace all above mentioned parts. I will outline details for the standard runs below.
Specs:
Agilent 1200 series w/ degasser and quad pump, Waters T3 250x10mm RP column + guard column, HPLC grade water and acetonitrile
Standard Method:
Sample: 6ul acetone and 0.5mg acetonaphthene in 1ml mobile phase
Injection volume: 2ul
Flow: 5ml/min
Column oven temp: 26C
Isocratic: 40% Water, 60% Acetonitrile
Equilibration time: 30 minutes
Runs:
9/5/12 - for acetonaphthene 19.35min @ 195bar
9/11/12 - 19.1min @ 210bar, 18.95min @ 208bar, 19.24min @ 204bar
9/12/12 - 19.02min @ 205.3bar, 19.1min @ 206.4bar
As you can see my pressure fluctuates slightly but is always very stable during equilibration and run, but my retention times are always 0.1min to 0.4min shifted earlier. My suspicion is the temperature in the room, because though my column is in a column oven, the cover cannot be inserted because the guard column holder is too large, thus half the column is exposed to the room temperature, which changes daily. The acetone peak elutes within the same time, though for today it was shifted about 0.05min earlier too. My boss believes it could be air trapped inside the column (from a single large air bubble entering earlier this week), but I've been running solvents at above 1500psi for hours, and see no blips in the detector. I've also backflushed the column in attempt to "purge" the bubble from the column (suggestion by boss) for an hour at 0.5ml/min MeOH.
Any suggestions will be appreciated, and I apologize for lots of text.