Shimadzu UPLC injector reproducibility problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi guys.
We bought two Nexera UHPLC about a year ago. Since then, we had several problems with reproducibility, for example, having RSD about 27-30% with some methods, last one were six injections, and the third one was about 1/3-1/4 of the area. Don´t know why. I know is very important needle wash, R0 solvent, purge, etc, but it`s more difficult than HPLC. Could anyone bring me some tips?
Q. F. Ignacio Viera
We had a similar problem with our Dionex system and this was traced to air bubbles in the autosampler syringe. We were caught out as the Dionex system was different design to our usual Agilent systems. If your system does use rinse solvents and a syringe design make sure the solvents are well degassed.

Unfortunatly I can't find any information on the Nexera autosampler design to see if this could be a problem.
Insanity: doing the same thing over and over again and expecting different results.
Albert Einstein, (attributed)
US (German-born) physicist (1879 - 1955)
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry