Hi Folks,

I am running an established method for OVI analysis and am noticing a "bonus" peak on every second injection in my sequence. Any thoughts?
Could this be carryover? It's a little hard to diagnose with the amount of data you provided. If you have 3 injections do you see the peak in the second and third injection. If you inject a blank immediately after the sample is the mysterious peak there? What happens if you extend the run?
Its a very sharp peak which makes me think it wouldn't be carry over
Have you observed your autosampler injecting? See anything unusual?

Can you show us a chromatogram, or list the method parameters of the ALS?

Your column description, the type of inlet liner, the installation parameters of the column you used and have checked are all useful pieces of information.

best wishes,

Rod
When was the last time injector maintenance was done?
5 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry