I have a pump pressure/flow rate problem with agilent 1100 binary pump. I found a similar post here and quite similar problem that with lower pressure (45 bar with 60%ACN/Water) than usual (83bar) and only 70% flow rate (0.7ml for 5 min with 0.2 ml/min setting). But without column it flows 100%. We have no visible leak anywhere.
So we changed outlet check value, active inlet cartridege, and pump seal - but still no success.
One thing we noticed is that with one pump pressure got lower 55 to 45, then the other pump also got lower to the same.
Any suggestion?
I would appreciate any help on this.
Margot-1 wrote:
I have an Agilent 1100 quaternary pump with a flow rate problem. The flow rate tests low ~0.7-0.8 mL/min when set at 1.0 mL/min. This was tested at the end of a column in a 10 mL grad cylinder after 10 minutes. The column pressure was ~220 bar and steadily rose to ~270 bar with 1:1 ACN:H2O mobile phase. Each of four channels (ABCD) tested the same. The pump has new check valve cartridges (inlet & outlet), a new purge valve filter and a new gradient mixing valve. I did not replace the active inlet valve that holds the inlet cartridge yet. The flow rate tests fine with a coil of peek tubing (without a column) at 1.0 mL/min at ~40 bar or at 5 mL/min at 170 bar. We had similar low flow rate results with a brand new column. Any ideas?