I clean up urine for nitrite analysis by C18 (Bond elut 100mg, 1mL) in "pass-through mode" as follows:-
1. activate column by 2x1mL MeOH, then 2x1mL H2O.
2. pass 1mL urine and discard the flow-through.
3. pass 0.3mL urine and collect the flow-through. When sample stops dripping, apply suction at </=5mmHg.
4. Then dilute 1 in 10 for nitrite analysis.

However, I can't obtain a linear curve with fortified standards. The responses are not precise. Is there anything wrong with my C18 cleanup procedure? Is there any benefit if I dilute the urine sample before loading to C18 since this may lower the viscosity and hopefully enhance reproducibility?

I appreciate your comments.
Probably you should wash the column with water. The 0.3 ml is absorbed by the column and does not elute even with vacuum.

Peter
Thanks for your comment.
I tried passing 1mL 1 in 10 diluted urine and discarding the flow-through. Then applied 0.6mL 1 in 10 diluted urine and collected the flow-through. The curve was linear and reproducible now.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry