So: I'm a relatively experienced GC operator, new to the worldof MS, and I'm having some issues.
I have a new ScionSQ from Bruker, and have been trying to get it up and running. I had some challenges, the original column install had about 6" of column touching the floor of the oven, so we had some major bleed issues. I thought I'd cleared that up (took a meter off the column, baseline looked better each run), but it appears to be back with a vengeance. I'm kind of thinking it never really went away.
I'm getting an AutoTune failure--"Not enough ions to tune." If I turn on the MS in scan mode, and turn cal gas on, I see a peak at m/z 66....and that's it. Any clues? Is my column totally messed up and not actually flowing to the MS?
Apologies in advance, as I said, I'm a noob, and may have some silly questions along the way to taming this beast.
Cheers,
GCMSN
ETA: Air/Water report looks...funky. 28/32 ratio 35:1, 28/18 350:1, the absolute size of the 28 peak is 1x10^7, it tells me below 9x10^7 is acceptable.