acetic acid buffer
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
9 posts
Page 1 of 1
i used hplc to detect metabolite products of dye decolorization. i used solvent acetonitrile and acetic acid buffer 30 mM pH 4.5. after i inject my sample i always find peak of acetic acid with high intensity. i just wonder, how to remove presence of acetic acid in peak? thank you.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Use 30 mM acetate in both your diluent and your mobile phase. If you *still* see a peak for acetic acid, that means that the excess is, in fact, part of your sample.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 10
- Joined: Wed Apr 28, 2010 1:01 am
thank you. i forgot to tell something. my sample also contains acetic acid buffer. when i inject only water, acetic acid did not appear. just right now i inject dye in water (not in acetic acid buffer), acetic acid peak appears again.
by the way, i little bit dont understand about addition acetic acid in diluent and in mobile phase. what's the meaning of diluent? thank you so much for your respond.
by the way, i little bit dont understand about addition acetic acid in diluent and in mobile phase. what's the meaning of diluent? thank you so much for your respond.
-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
When there's really no acetic acid in your sample solvent, there are two possibilities: Either the peak you're seeing IS acetic acid and it is comming from your dye sample (and not the solvent). Or it is NOT acetic acid but some other component of your dye sample.when i inject only water, acetic acid did not appear. just right now i inject dye in water (not in acetic acid buffer), acetic acid peak appears again.
Just to be sure, that peak is positive (upwards), right?
If possible, you should use your mobile phase as solvent for your sample. This way, you minimize the possibility of seeing extra peaks that do not originate from the sample.
-
- Posts: 10
- Joined: Wed Apr 28, 2010 1:01 am
hello.. i change the wavelength and acetic acid peak has already dissappear. however, i still confuse about how to choose wavelength. my initial sample has maximum absorbance at 595 nm. so when i inject sample in that absorbance, the initial sample peak is appear. but my treated sample (metabolite product from my initial sample) can not detected. so what should i do? i should get the new absorbance to detect other compounds? sorry, too many question. it's the first time i use hplc. thank you so much..
-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
Generally speaking, you should set the wavelength so you can detect all analytes of interest with satisfying sensitivity.
Do you want to see the initial substance AND its metabolite? Do you want to see only the metabolite? Do you want to see only the initial substance?
If you want to analyse only one of these two, you should set the detection wavelength to its absorption maximum. This way, the other peak might get very small or even dissapear completely, but who cares - you're not interested in it, right?
If you want to analyse both simultaneously, you'll have to find a wavelength where both show enough absorption. If that's not possible you can use different wavelengths for both analytes - either in one run, if your equipment (detector) is capable of it (i.e. simultanous measurement of two wavelengths or changing of the wavelength between the two peaks), or you'll have to split the analysis in two runs with different wavelengths.
Do you want to see the initial substance AND its metabolite? Do you want to see only the metabolite? Do you want to see only the initial substance?
If you want to analyse only one of these two, you should set the detection wavelength to its absorption maximum. This way, the other peak might get very small or even dissapear completely, but who cares - you're not interested in it, right?
If you want to analyse both simultaneously, you'll have to find a wavelength where both show enough absorption. If that's not possible you can use different wavelengths for both analytes - either in one run, if your equipment (detector) is capable of it (i.e. simultanous measurement of two wavelengths or changing of the wavelength between the two peaks), or you'll have to split the analysis in two runs with different wavelengths.
-
- Posts: 10
- Joined: Wed Apr 28, 2010 1:01 am
actually i have to get all compounds (initial and its products)
and i got difficulties about it. yeah, i think i should repeat them. i use fraction collector to collect the samples.
-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
Ah, so you're doing preparative chromatography. Bad luck - reruns are a bit more of work
I'd suggest to try a low wavelength, something like 210nm - should increase the chance to see everything. But be careful, you'll see more junk, also. You should develop the method on an analytical system first before going to preparative work...
I'd suggest to try a low wavelength, something like 210nm - should increase the chance to see everything. But be careful, you'll see more junk, also. You should develop the method on an analytical system first before going to preparative work...
-
- Posts: 106
- Joined: Wed Nov 23, 2011 11:35 am
hellohello.. i change the wavelength and acetic acid peak has already dissappear. however, i still confuse about how to choose wavelength. my initial sample has maximum absorbance at 595 nm. so when i inject sample in that absorbance, the initial sample peak is appear. but my treated sample (metabolite product from my initial sample) can not detected. so what should i do? i should get the new absorbance to detect other compounds? sorry, too many question. it's the first time i use hplc. thank you so much..
595 nm is the visible range, that's why you may easily see your initial peak (which is a dye). As far as i understand your metabolite is decolorized which means your metabolite has less conjugation then initial product. This causes the wavelenth shift to lower (below 350-400 nm). In that case you should select two wavelenghts to detect both compounds.
for acetic acid case; maybe acetic acid forms during the chemical transition. Not to detect acetic acid you may select a wavelenth higher than 220-230nm if possible
good luck
9 posts
Page 1 of 1
Who is online
In total there are 108 users online :: 1 registered, 0 hidden and 107 guests (based on users active over the past 5 minutes)
Most users ever online was 4404 on Wed Nov 05, 2025 8:47 pm
Users browsing this forum: Baidu [Spider] and 107 guests
Most users ever online was 4404 on Wed Nov 05, 2025 8:47 pm
Users browsing this forum: Baidu [Spider] and 107 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science