To shorten the run, can I use 2mL/min flow to fast flush for 7min, then run with 1mL/min till the peak comes out?
Thanks in advance for your comment.
Advertisement
Discussions about IC and related topics
How so?I worry if I use 2mL/min all along, the interference peak around the target peak will be too close to it and peaks will not be sufficiently separated.
Don't worry and just try. You will see if it's possible...I worry if I use 2mL/min all along, the interference peak around the target peak will be too close to it and peaks will not be sufficiently separated.
For UV-detection? Yes.Thanks for your all your comments.
Another point, doubling the flow rate means decreasing peak area by half?
Yes I use UV detection, so I cannot increase the flow rate and sacrifice sensitivity around my target RT.For UV-detection? Yes.Thanks for your all your comments.
Another point, doubling the flow rate means decreasing peak area by half?
It's been done. The major shortcoming is that peak area is essentially the product of absorbance and residence time in the flow cell. Changing the flow adds error to the "residence time" part . Benefit? Not much. As I indicated in my earlier post, today's columns really have a fairly shallow efficiency vs flow curve (at least for small molecules; biopolymers are a different story), so you don't gain much by slowing the flow at the beginning of the run. In our method development courses, i tell people to set the flow rate as high as possible consistent with their level of paranoia about pressure, baseline noise, and column lifetime.What are the shortcomings or benefit (if any) of flow gradient?