Shimadzu prominence data reprocess problem

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Hey guys. Here´s my problem.
I have a sample with two peaks. I have to integrate them as a group, and quantitate them vs an external standard, which has the same two peaks. I did everything the manual says to do (a very short explanation I must say), beleave I had read it all, but it didn´t worked. The software (LC solution) takes the area from one peak, not the sum of two.
I gave up. Need some help. Anyone knows hoe to do this?
Q. F. Ignacio Viera
Try to change the peak integration parameters!

Hope I´ve helped!
Just did this for a customer. Two peaks within the retention time window specified, integrated separately and summed as one compund Need to select to include all peaks on the Quantitative tab of the Method.
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry