GC/MS possible leak
Posted: Thu Jun 14, 2012 4:49 pm
What is the typical difference in vacuum between isolated MSD vs. MSD connected to GC/MS?
When I block the transfer line I am getting very good vacuum 6x10-7 When I connect the column, I am only getting 4x10-5 with He set at 2PSI. No apparent leaks when looking at the spectrum (14, 28, 18, 32 are all very low compared to helium). It bothers me cause the sensitivity is very low so it doesn’t even tune sometimes. This is Perkin Elmer Clarus instrument, unlike chemstaiton I do not believe there is a way to do full scan or air/water scan.
When I block the transfer line I am getting very good vacuum 6x10-7 When I connect the column, I am only getting 4x10-5 with He set at 2PSI. No apparent leaks when looking at the spectrum (14, 28, 18, 32 are all very low compared to helium). It bothers me cause the sensitivity is very low so it doesn’t even tune sometimes. This is Perkin Elmer Clarus instrument, unlike chemstaiton I do not believe there is a way to do full scan or air/water scan.