What is the typical difference in vacuum between isolated MSD vs. MSD connected to GC/MS?
When I block the transfer line I am getting very good vacuum 6x10-7 When I connect the column, I am only getting 4x10-5 with He set at 2PSI. No apparent leaks when looking at the spectrum (14, 28, 18, 32 are all very low compared to helium). It bothers me cause the sensitivity is very low so it doesn’t even tune sometimes. This is Perkin Elmer Clarus instrument, unlike chemstaiton I do not believe there is a way to do full scan or air/water scan.
Those pressure differences sound correct to me, assuming the Clarius pumps are what comparable to a low end Agilent GC/MS.

Some MS will have greater pumping capacity (multiple pumps, etc), but the entry level ones I am used to will read in that range with a 1 mL/min carrier flow.
2 posts Page 1 of 1

Who is online

In total there are 36 users online :: 1 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry