Quick question, I have a method setup on ESI +, I am interested to investigate ESI - , so when switching from ESI+ to ESI- does the original mobile phase need to be changed, or can that be kept the same, and what will happen to signal intensity. My Mobile phase conditions are
A) 10mM Aqueous Ammonium Acetate pH 3.0, and B) ACN.
Since the parent ion will presumably change to MW-1, is a quick infusion necessary with a MS2 full scan, or would a product ion scan of MW-1 be optimal as a quicker approach to method development.
Is signal intensity typical lost for negative mode? I know that most methods are using ESI+ , what is the key advantage of using + over -.
I have some strange analyte behavior in + mode so I am searching - mode to see if the trend follows.
Thanks Kindly
Funny things can happen in the electrospray ionization mechanism, and "wrong-way-round" ionization has been widely reported - you can have an acidic mobile phase and still get [M - H]- negative ions. If you don't see it when you try at pH 3, by all means try again around pH 7.
In positive mode the acidity in the highly charged spray is significantly greater than the original mobile phase. I would thus assume the opposite can happen in negative mode.
I think there are just a lot more analytes that respond well in positive mode, and even ones that respond better in negative are often still highly visible in positive. In the charged spray environment, many carbonyl groups can be easily protonated that would not normally do so in solution. But in negative mode it's still a challenge to pry off a not-very-acidic proton. Of course in negative mode there are also adducts that are sometimes the best visible ion.