issue : Retention time shifting in ion chromatography
10 posts Page 1 of 1
instrument: metrohm
column : metrosep A supp1 250
mobile phase : 0.318 gm in 1000mL water
detecion: sulfate
initial Rt :24(very 1st injection) and after 100 injections its 21
can any one suggest any solution for it
thanks in advance
With Warm Regards
Pr. D.Sudhir Kumar M.Pharm
Research Associate-IV
AUROBINDO PHARMA
is the RT 24 throughout the first 100 injections and then suddenly 21 min at 100? Or is it a gradual change from 24 min to 21 throughout 100 samples? Does it go back to 24 on its own or do you have to do any maintenance (e.g. backflush column etc.).
~Ty~
Through out the analysis/or till the 100 inj(its not the exact figure) runtime was 24 and for the next time when i performed the analysis its 21 I tried with diff systems, diff M.phase, at-last when i tried with new column it was again 24
and one more thing is that the old column which i was getting 21 reduced its Rt.
With Warm Regards
AUROBINDO PHARMA
How long did the column sit between those sets of runs? what was the storage buffer (and what did the manufacturer recommend). It looks like your column is dying during storage.
the mobile phase which i run is sodium carbonate anhydrous(0.3 gms in 1000mL water) buffer, and its the storage solution recommended by the column vendor
With Warm Regards
AUROBINDO PHARMA
All the symptoms point to the column deteriorating, so the next step would be to contact the vendor.
what is the sample matrix?
back pressure on the column (initially and with reduced RT?
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
yes the pressure is gradually increasing
initial - 7.8
present - 9.4
With Warm Regards
AUROBINDO PHARMA
the sample matrix is enoxaparin injection
With Warm Regards
AUROBINDO PHARMA
Are you using a Guard column?
If yes, how is the back pressure without Guard?
Did you replace the guard column?
If no, check with Metrohm India about column regeneration on this application.
Dr. Markus Laeubli
(retired)
Metrohm AG
9101 Herisau
Switzerland