by
GHGirl » Wed Aug 01, 2012 9:40 pm
I thought I'd post a quick update for others who may run into the same issue in the future.
I've been working closely with Agilent to find a solution to running three detectors simultaneously on our 7890A Gas Chromatograph. One of the major challenges is that our particular system is capillary, including the FID.
As Gasman mentioned, the kit that Agilent offers for running a TCD and FID in series is only for a packed FID, so that wasn't an option for us. But we did manage to clip the 1/8" flare off the TCD vent and connect the 1/16" portion of the TCD vent to a 0.53mm capillary using a Swagelok union, which could then be inserted into the FID (as Rod suggested). In fact, we tried several variations of this connection...
Ultimately, the TCD mechanics (switching from the carrier to a reference) was a problem for the FID baseline, and we could never get the noise quiet enough to detect trace amounts (i.e., atmospheric levels) of methane on the FID. So for now, running the TCD and FID in serious on a pure capillary 7890A system does not seem to be an option.
Our solution to running three detectors simultaneously? We connected the column (we switched to an HP-PLOT/Q) to a Deans Switch, with the OFF position running to a Two-Way Split and the TCD and FID (so the sample is split in half), and the ON position running directly the the ECD. In this way, we can time the Deans Switch so our CH4 and CO2 reach the TCD and FID, and the N2O reaches the ECD. We can also run C2H2 (see my post, Acetylene Poisoning ECD - Fact or Fiction?) to the TCD and FID, regardless of whether C2H2 is a true problem for ECDs.
As always, thanks for help!
Greenhouse Gas Girl