Page 1 of 1

Empower 2 running Shimadzu HPLC

Posted: Thu May 31, 2012 2:54 am
by Brockboy02
Hey everyone,

Anybody out there that has experience running Shimadzu HPLC' s (Prominence) using Empower 2? I am a bit perplexed at which gradient curve (pump B Curve) to use. I want a linear curve and thought that entering 6 would give me a linear curve. The resultant chromatography tells me it is definitely NOT linear. Maybe that is only true using Empower 2 to control Waters HPLC's?
Any help would be great!

Thanks!

Re: Empower 2 running Shimadzu HPLC

Posted: Thu May 31, 2012 3:01 am
by DR
Check the pump or controller manual - I'm pretty sure you need a 1 in there, not a 6.

Re: Empower 2 running Shimadzu HPLC

Posted: Thu May 31, 2012 8:25 pm
by JGK
6 is the linear gradient No if you are using Waters equipment

Re: Empower 2 running Shimadzu HPLC

Posted: Thu May 31, 2012 11:02 pm
by Brockboy02
Problem solved! The proper curve to use is 0. Shimadzu allows for a curve value from -10 to 10 (thats a lot of different curve types!), so one would assume 0 would be linear. And it is. Same as if ClassVP was controlling the HPLC. The curve of 6 IS only linear for empower controlled Waters HPLC's. Since the driver for the software is released by Shimadzu, this makes sense. I wonder, if the linear curve value would be 6 if a Shimadzu software CDS was used to control a Water's HPLC? Hmmmmmm......

Re: Empower 2 running Shimadzu HPLC

Posted: Fri Jun 01, 2012 3:23 pm
by DR
Ah - I was close, anyway. :)