We are experiencing numerous cases where we are unable to achieve the 8500 required for the Chemical Performance test according to the Agilent OQ/PV protocol. Has anyone else experience this issue and any resolutions?
-
- Posts: 158
- Joined: Wed Mar 05, 2008 2:56 pm
Hi
Firstly which inlet / detector setup do you have.
is the system leak tight
are you using an A/S, is the syringe ok, is the software configured correctly for the syringe
is this normal column or restriction column
are you using chemstation
is the liner correct
has the gold seal been replaced
Firstly which inlet / detector setup do you have.
is the system leak tight
are you using an A/S, is the syringe ok, is the software configured correctly for the syringe
is this normal column or restriction column
are you using chemstation
is the liner correct
has the gold seal been replaced
-
- Posts: 3
- Joined: Thu Dec 29, 2011 3:03 pm
Hello,
6890 FID , split/splitless inlet
restriction column
using a 5ul syringe, software configured
leak test passed
using chemstation
universal liner being used
Gold seal was checked and was replaced
6890 FID , split/splitless inlet
restriction column
using a 5ul syringe, software configured
leak test passed
using chemstation
universal liner being used
Gold seal was checked and was replaced
-
- Posts: 158
- Joined: Wed Mar 05, 2008 2:56 pm
Hi
If the above are correct then perhaps check the FID flows, should not make too much difference though.
Also double check the syringe configuration is set to 5ul rather than 10ul.
Sometimes it is worth carrying out the leak check with the column installed, i.e. push a septa onto the column end.
Also check the split flow is correct.
good luck
If the above are correct then perhaps check the FID flows, should not make too much difference though.
Also double check the syringe configuration is set to 5ul rather than 10ul.
Sometimes it is worth carrying out the leak check with the column installed, i.e. push a septa onto the column end.
Also check the split flow is correct.
good luck
-
- Posts: 3
- Joined: Thu Dec 29, 2011 3:03 pm
Thanks, will do the checks and see what happens.
5 posts
Page 1 of 1
Who is online
In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 32 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science