The interesting part is the the QL of the impurities improved by a factor 3. The reason for this is that the detector noise of the Allaince system is 10 fold lower compared to the Acquity system (which was requalified some weeks ago).
Which detector rates/speed have you used with both systems?
- The weak needle wash has a huge impact on the chromatography (this is not written anywhere in the manuals, and not even the technicians knows about this!)
Amazingly this seems to be true, e.g. in the Acquity UPLC System - Operator#s Gide (7150082502/Revision C) it seems that this well-known issue is mentionend only once (Chapter 'Diagnostics and Troubleshooting':
To stop a needle wash routine before it finishes:
...
Caution: Do not abort the sample needle wash sequence. Doing so may leave strong solvent in the sample needle which can adversely affect chromatography."
Maybe a more detailed introduction of the autosampler in the manuals should be done by waters. During the launching of the UPLC, one of the most important topics was always the functioning of this novel autosampler and the requirements (e.g. weak needle wash should have the same compressibilty as the sample solvent).
- The detector show much higher noise levels that ordinary UV-detectors. You loose all the sensitivity that you gained by the higher peaks
I can't believe that this is in general true. According to my experience, most of the problems with noise using a UPLC is caused by a wrong detection rate or problems with mixing of highly uv-active mobile phases.
- The column oven cannot take more than 150 mm columns
I think Waters offers now also special column ovens for this problem which design looks accidently like as known from the Alliance
Also since some years, there is finally a real usefull solvent tray for the old-fashioned Alliance availible, like as known from the UPLC
But in general, doing real sub-2-micron chromatography, there should usually be no need for such long columns. Best starting point in my opinion is a 5 cm length
(and sometimes it may be also to long) and only for special applications a longer column may be considered.