Hi all
Im usin μBondapak CN Waters column for analysis of BenzCl and im askin you any tip for cleaning and regeneration of this type column.
Thnx in advance
That is an *old* column technology! Best to check with Waters for any specific recommendations, but in general do a reverse-flush with something that will solubilize whatever flavor of CRUD is on there.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
CN columns are pretty tolerant of solvent- you can wash with organic solvent or water depending on what the "crud" dissolves in. You may need to go thought an intermediate solvent (like isopropanol) to go between normal phase solvents (like hexane) and water. Being silica based, you want to keep the pH below 7.5.
Im usin a combination of 1-hexane sulfonate sodium salt with ACN (55:45) for my analysis. The retention times of peaks are not stable and i have some problems with the cleaning and make the column usable again. I usually wash the column with water and store in ACN:H20 80:20. Can you give any advice for better results?
Thnx in advance
Let me elaborate on my earlier post. You have to make a judgement about what type of contamination (CRUD = Chromatographically Retained Undesirable Debris) is likely to be in your samples, and then flush the column with something which is a good solvent for that type of contamination. What you are doing sounds fine if the contaminants are salts, but would be totally inappropriate if you have fatty or non-polar CRUD (100% ACN or even ACN with some dichloromethane would be better in that case).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thnx for the help, we will continue the procedure we follow, maybe we need to replace our method with a new one usin a C18 or usin a brand new CN column after an appropriate period
I just started using a Cn column and in the instructions it says to store in IPA.
7 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry