Area counts rising for subsequent injections

Discussions about GC and other "gas phase" separation techniques.

7 posts Page 1 of 1
For the sake of simplicity, let's pretend I have a sample with two components: analyte and solvent. What are common causes for the area count of the analyte to increase between runs when injecting from the same sample vial?
Analyte carry over somewhere in the chromatographic system.
For the sake of simplicity, let's pretend I have a sample with two components: analyte and solvent. What are common causes for the area count of the analyte to increase between runs when injecting from the same sample vial?
For the sake of simplicity :wink: most obvious reason is evaporation of solvent (from vials - provided it is more volatile than analyte).
Actives sites in the sample flow path which become blocked with the analyte after continued injections.

These active sites could also be sites of catalytic decomposition of the analyes as well. So if they become 'consumed' then more of the analyte peak will elute to the injector.

Of course, all this assumes that your detector is not changing, becoming more sensitive due to some cause.

Murphy is alive and well and has been known to play all sorts of tricks on the unwary.

So be wary.

best wishes,

Rod
If the most simple possible scenario: If you have two runs there is an exactly 50% chance that the peak area for the second run will be bigger than the peak area for the first, and there is absolutley nothing that you can do about it.

Peter
Peter Apps
Peter's reply is accurate to within one part per infinity, which is pretty close as far as accuracy goes.

It is possible that the two would exactly match.

But that likelihood is like having two posters exactly agree on an answer, which is also one part per infinity, give or take a part.

best wishes,

Rod
What's the detector?
7 posts Page 1 of 1

Who is online

In total there are 34 users online :: 1 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry