Default 6890 FID cutoff current is 2 pA.When dry air source was changed, I've got a fault 214 "flame out". It tries to igniting two times then shut down H2 and dry air. Has anyone experienced this situation? The machine is agilent 6980
Thank you for your answers and helps...
it sounds so logical. I will try tomorrow.Default 6890 FID cutoff current is 2 pA.When dry air source was changed, I've got a fault 214 "flame out". It tries to igniting two times then shut down H2 and dry air. Has anyone experienced this situation? The machine is agilent 6980
Thank you for your answers and helps...
If your air is much cleaner than previous air source, signal current may be lower than cutoff setting in your 6890.
Chech signal level when FID is lit and compare it with cutoff setting.
Thanks for your response. I will check the H2 response now. And I uploaded the response of GC on youtube. I changed the lit offset from 0.1 to 4 but it doesnt work.Do you have a flow meter? Measure the flows individually and make sure they are correct. Try blowing gently on the ignitor when it is glowing. You may also want to turn the hydrogen up a bit for a minute or so before trying to light. If it still isnt working try using a cigarette lighter. Once lit make sure the signal is normal (not too high > 15 DU and not too low.. I've seen situations where the fid is plugged a bit with graphite and it was difficult to light and once lit the signal was very low around 2-2.5 pA and would easily go out. Flows measured correctly, but when I removed the jet and column and held a flashlight underneath I could only see a tiny bit of light.
Another good way to check for a restriction in the jet is to turn all gasses off and then turn the H2 on and set it for a flow of 90, scroll down and check the reading of the make up flow (even though it is off.) If there is a plug some of the H2 will go backward into the pressure controller and be detected as make up flow, a normal reading would be 0-3 and a plug could show like 9-15 mL/min.