Hi guys.

I have some questions about referenece wavelength, talking about DAD modules from Prominence and HP 1100.

First of all, some basics that I haven´t got really clear. When is it really necessary using it? Because I know when your solvents are meesing up your peak, or when there´s something in your sample matrix messing too.

And why 1100 let´s D2 lamp be reference until 400 nm, and Prominence 600 nm (because if you put a reference under 600, and turns on VIS lamp, gives you trouble)? There are 200 nm of difference. The range is wider. Don´t know if it is on the lamp technology or detector technology.
Q. F. Ignacio Viera
The generally accepted wavelength range of analytical utility for a deuterium lamp is 190-400nm(some say 360nm) since the UV energy drops to about 10% of the peak power at ~400nm.

Use of a reference wavelength rarely helps and is (in my understanding) primarily intended to correct for changes in refractive index of the mobile phase.
A. Carl Sanchez
2 posts Page 1 of 1

Who is online

In total there are 28 users online :: 2 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry