I am running an LC-MS/MS method using a gradient with water and acetonitrile (both with 2 mM ammoniumformate and 0.2% formic acid) on a PFP propyl column, 50*2.1 mm, 5 µm with guard column 10*2.1 mm, 5 µm.
I have a problem: after each run, the (equilibrated) pressure increases with 10-20 bar.
The problem occurs even when the sample is just water (which works fine on our other LC-MS/MS), so the sample is not the problem.
If I backflush the column when pressure is increased, the pressure returns to normal. If I start my analyses again, the pressure again rises. The guard column alone has always the same pressure, so the increased pressure is a problem of the column!! This is confirmed since backflushing the column with or without the guard column solves the problem.
Has anybody an idea how I can fix this, since it is not very convenient to backflush the column after every 5 runs or so. And I suppose this is not ideal for the column lifetime...
Thanks for your help!