Hello

I've ten years experience in IC included my degree on validation of IC methods.
I've experienced a crytical problem with my new ICS1100 running with AS9HC & carbonate 9mM eluent.
i try to describe the situation.
the system is new, it worked fine for a few months but now i experienced a drammatic decrease in retention times of all anion peaks.
It starts with change in resolution of sulphate/Posphate... continues with sovrapposition of Nitrate with sulphate+posphate & ends with death of column.
I use the same eluent & water since 10 years.
Water is bi-distilled & filtered with conductivity usually < 1uScm-1.
Carbonate is (i wish) as pure as possible.
Contaminants that kills colums that i know are:

-CrVI (it's absent i've controlled it AA it's >LOD in flame i can control in A.A./E.T.A. but i want to exclude ppB levels i don't believe it's present

- grease (the system is completly new ics 1100 it mustn't have problems with grease i wish

- all the lines for eluents are inert

- i've checked eluent for presence of fe3+ but it's < LOD in flame

- i've thinked about free Cl2 in water but it's bidistilled water...
& i don't know what's the effect of cl2 on PSDVB polymers

i don't know what to look for.

It kills AS9-HC & i've fear that it contaminate the system!!!!!
What should it be?

I'm desperate i'll request a dionex visit but i don't know what's the problem.

The eluent is prepared with the same procedure (column data sheet) since ten years: mother solution .... diluition to 9mM carbonate.... etc....

HELP ME.
Treat yourself to a bottle of HPLC-grade water, or at the very least, a milli Q apparatus.

I would pump 40 mM Na2EDTA through your system (and column, but not detector) o/n at a low flow rate. Follow this with water, then several column volumes 0.5 M NaClO4 20 % MeCN, followed by water, then re-equilibrate with your mobile phase.

Consider using a cation exchange guard, or, better yet, a WCX-SAX mixed-bed guard. Exogenous Fe+++ is known to bind strongly to SAX columns, deteriorating performance. The WCX guard will mop up Fe+++ in the mobile phase, preventing it from fouling the analytical column.
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry