Fronting peaks in GC

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
The analyte I'm interested in fronts quite badly.
The compond is a primary amine and I'm running on an RTX amine phase.

Loading seems OK for the column I'm using (30m x .32 x 0.5u) and I think I've got correct flows etc.

Any ideas whats causing this and any tips to try and avoid it ?

TIA

Stanlee

Fronting may be due to column overloading. Try to inject less or increase the column capacity by increasing the film thickness.

as Stanlee suggests: a good thing to try whenever peak shapes are poor is to inject less sample (injection volume and/or concentration) to see if the shape improves.
Another common cause for fronting is trying to elute a compound at too low a temperature. The earlier posts I wholeheartedly agree with as well.
try setting your initial temp 5 degrees below the boiling point of your solvent.
Ex Mecl2 BP is 40C, Set temp initial at 35C
5 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry