aidnai » Fri Mar 02, 2012 10:56 pm
Hi all,
Thanks again for the replies.
AICMM -- your suggestion succeeded in lowering my signal, but not my noise >_<
I have a .53mm column x 30m, so I'm running it at 5mL/min with a head pressure of only about 3psig. I wanted to keep the make up low as possible but not get into situations where there is unswept dead space (my make up is about 2mL/min) ... I'm used to packed columns myself so the numbers feel weird to me but I was going off of agilent guides while choosing rates/pressures.
I really believe the problem is electronic in nature, as evidenced by the drastic reduction in noise when the oven/oven fan is turned off.
Aldehyde -- Thanks for looking into things for me. I did feel like the noise should be lower, you pretty much confirmed it. I did the noise determination you recomended and got the following results

My scale is still 'counts' whatever the hell that is, instead of 25uV or something I can understand. As close as I can tell from my printouts, 1000mV is equivalent to 200000 counts, giving a conversion of 25uV -> 5 counts. That would (correct me if I'm wrong) make your PtoP noise 1.1675 counts vs. my 270.9896 counts ...
I checked the column, it isn't touching anything except its cage (I kinda think you had a packed column in mind anyway). I also checked the baseline with the detector off -- dead flatline. While probing around, I found that some of the TCD filament/temp sensor leads are losing their insulation - i can actually see bare wire on a few. Should I consider re-wrapping them?
AP