Hollow » Thu Mar 01, 2012 9:12 pm
I don't see the test as "useless" because it tells you that nothing seems to be wrong within the system nor with your water nor ACN.
So it is probable that there is something with the ion-pair salt and/or your phosphoric acid.
Have you tried to run the blank gradient with A=only the salt in water and A=only phosphoric acid in water?
And another question:
How do you prepare your eluent? Is there anything involved that could contaminate your mobile phase, e.g. pH-probe?
Again, the heptansulfonic acid salt is not just a buffer, but acting as ion-pair reagent.
It will stick to your C18 surface with it's C7-tailc turning your column "temporarly" into a cation exhanger by the sulfonat group.
The concentration of the equilibrium between C7 and C18 depends on temperature and the composition of the mobile phase.
If the ACN concentration gets higher, the C7-load will decrease and be washed out. That's what the humps probably are.
Therefore it should not be used with gradients.
Have you already tried to use an ordinary phosphate buffer (Na or K, e.g. 10mM) at the same pH? Will the separation work?
If you really need this ion-pair mechanism, switch to isocratic elution (if possible), try a dedicated SCX-column,
where the sulfonic acid groups are fixed to the surface or stationary phase, or maybe you could try a mixed-mode column (a certain user here will doubtlessly help you on this

)
And yeah, I don't know the PE 200 so I can't give you any hardware specific help.