Lipid Area Reproducibility

Posted: Thu Feb 16, 2012 3:52 pm
by nick22
I am analyzing a mixture of three lipids by HPLC with ELSD detection. I get good separation on a C-18 column but the %RSD for one of the components is large (~15%) but the other two lipids show good reproduciblity (%RSD < 2%). This component is a PEGylated lipid and is the second peak to elute.

The problem is not carryover from a previous injection because injecting a blank after the sample results in no peaks in the blank injection.

Is this a problem inherent to using ELSD detection with PEGylated compounds?

Re: Lipid Area Reproducibility

Posted: Sat Feb 18, 2012 9:25 pm
by unmgvar
is it the only one with PEG?
if yes then
one possibility is of your procedure of doing the pegylation
how do you know that it is going well?

Re: Lipid Area Reproducibility

Posted: Sun Feb 19, 2012 3:14 am
by carls
PEGs are sticky. Try adding a small amount (10ppm) of glycerin to your sample

Re: Lipid Area Reproducibility

Posted: Tue Feb 21, 2012 7:22 pm
by nick22
Carls, thanks for the idea, I will need to order glycerin but will do so and will report back.

Unmgvar, we purchase the PEG lipid so if it is variable pegylation, we will need to find an alternate supplier.