Lipid Area Reproducibility

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
I am analyzing a mixture of three lipids by HPLC with ELSD detection. I get good separation on a C-18 column but the %RSD for one of the components is large (~15%) but the other two lipids show good reproduciblity (%RSD < 2%). This component is a PEGylated lipid and is the second peak to elute.

The problem is not carryover from a previous injection because injecting a blank after the sample results in no peaks in the blank injection.

Is this a problem inherent to using ELSD detection with PEGylated compounds?
is it the only one with PEG?
if yes then
one possibility is of your procedure of doing the pegylation
how do you know that it is going well?
PEGs are sticky. Try adding a small amount (10ppm) of glycerin to your sample
A. Carl Sanchez
Carls, thanks for the idea, I will need to order glycerin but will do so and will report back.

Unmgvar, we purchase the PEG lipid so if it is variable pegylation, we will need to find an alternate supplier.
4 posts Page 1 of 1

Who is online

In total there are 30 users online :: 2 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry