Hi,

We are trying for make naphthalene solution of 20ppm in DCM. When we analyse it in GCMS the conc. shown is 30% less around (14ppm).
This is consistent with several other solutions prepared with different solvents. Is this due to problem with GCMS?

Can it be fixed using internal stds?

Thanks
Jason
I'd use a solvent less volatile than DCM.

or use HPLC
Hi,

We are trying for make naphthalene solution of 20ppm in DCM. When we analyse it in GCMS the conc. shown is 30% less around (14ppm).
This is consistent with several other solutions prepared with different solvents. Is this due to problem with GCMS?

Can it be fixed using internal stds?

Thanks
Jason
How do you know that the concentration is "14 ppm" ?

Please do us all and yourself a favour and use proper units - the density of DCM is 1.3 g/ml, so if you are mixing up volumes and masses in calculating your "ppm" the problem disappears.

Peter
Peter Apps
3 posts Page 1 of 1

Who is online

In total there are 53 users online :: 2 registered, 0 hidden and 51 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 51 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry