GasMan » Thu Jan 26, 2012 1:28 pm
The only way that I can think of, is to have a free running program that will start your GC and also enable you to control the 9th valve via a time table.
You should also look at your present system. For example, if you have two 6 port gas sampling valves which inject at the same time, you could replace this with a 10 port valve that can be configured as a dual gas sampling valve. This would then free up a valve position. If you have a system with two liquid sampling valves, Valco do produce a dual LSV, again freeing up a valve position.
Another way to do the same thing if you have two valves actuating at the same time, is to 'tee' off the air supply to two actuators. By this I mean that you have the air supply to an actuator coming from a solenoid valve. You split the air line to two actuators, so that when you turn a valve on using the 'Valve X On' command, two valves will actually turn.
Regarding your thoughts on using a macro: user written macros will not run in ChemStation while the GC is running and collecting data. So as I see it, this would rule out you controlling a valve during a run.
One question. What are you using the Valve 7 and 8 positions to control?
Gasman