Missing Ion data

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
Hi there,

When I analyse my data I'm getting the message that I miss Ion's, e.g. ions 252 and 127.

I do a full scan and a SIM within one run on a Agilent 5973 MSD.Chromatographic nothing is wrong, the peaks are well seperated. I still can intergrate and quantify them. But I see this weird message.

Does anyone knows what to do?

Thansks in advance.

Regards,

Wouter
get a scan of one of your peaks then in the command line type in tab then hit enter. Put the exact mass of your sim compound in the method ( ie 202.4).

Whoops just read your complete message! 5973 can not do sim/scan! The software leads you to believe it can be done. Even if you have had an outside source reconfigure your 5973 to do so, Agilent will not support it ( want to sell 5975's).

You can do what I do, run 2 methods/sample. One sim and one scan. Kinda a hassle but they won't buy me a 75!
Hi Bigbear

When we bought this MS, it is a second hand, they told us that is was upgraded to 5975, or something close to it. I've done a lot of testing in the meanwhile and this machine can do sim/scan in one run. Gladly because one run takes 50 min. For some of my pesticides I analyze them both sim and scan. And when it's not visible on Scan it still is on SIM, Very nice in fact.

It's not that this is bothering me. But every now and then I notice it and than I'm wondering what is going one. But I will give it a try what you mentioned.

Regards,

Wouter
Regards,

Wouter
That's good that you have the upgrade. You will have to know the exact mass for all the compounds you will use durring the run. I got the exact mass for my instrument by running a sim method and acquiring data 0.1 D apart. So for example if my mass was a nominal 282 I set up a group to get all masses from 181.7--182.6. After the run I would take the avg. of the peak and tabulate. The mass with the highest area would be what is entered into the acquisition method. Then set the dwell time so you get 15-20 scans across a low concentration peak. It is quite a bit of work, but I was told not to trust the apparent exact mass data from a scan run.

Good luck
Hi Bigbear,

That is a nice solution when you have little components. I'm running with 170 components, and increasing :wink:, so this is not really a solution.

It is not that it is bothering me, but I'm wondering why this happens? What causes my MS to do so?

Anyone suggestions?

Regards

Wouter
Regards,

Wouter
Yes I've seen it to. As I said you need the exact mass of the compound in the SIM method and that same mass needs to be entered into the data aalysis method.
6 posts Page 1 of 1

Who is online

In total there are 37 users online :: 2 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot], Semrush [Bot] and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry