DR » Tue May 03, 2005 5:24 pm
It is something eluting from the column, but the source is not the column.
Try this - run at your initial conditions for ~20 minutes, then run the gradient twice without actually injecting anything. If the peak is larger in the first of these chromatograms, then the source is probably either the water used to prepare the mobile phase, or in the salt (less likely if you're using LC grade salt). The peak would be larger in the forst chromatogram because you have allowed more of it to build up on the column than in the second run.
To eliminate the contaminant (usually something from the water polishing system's resin bed - styrene monomers, dimers) you can:
-try bottled LC grade water
-install a small C-18 column between the A pump and mixing chamber if you're running a 2 pump system (remember to clean this occasionally)
-use a Empore Extraction disk to pretreat your water before making the A phase (a 3M product available from Varian & probably other vendors). Empore disks are available in a 47mm size to fit common filtration set-ups. You filter a liter or two of water, rinse the disk with MeOH (catching that in something other than your filtered water container), repeat. This is the best long-term solution to the problem that has plagued gradient chromatographers, especially at lower wave lengths.