Resolution of Bis(2-Chloroethyl)ether and Aniline
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
2 posts
Page 1 of 1
Can anyone help me get resolution between Bis(2-Chloroethyl)ether and Aniline?
-
- Posts: 310
- Joined: Thu Jun 18, 2009 7:59 pm
Yeah, it's not great. We get baseline (barely), using a 20 m x 0.18 mm 0.18 um Restek RXi-5Sil-MS, starting at 50 C. It's tricky - you need to balance your gas flows well, so that you're at optimum flow throughout the front part of the chromatogram.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
Laboratory Director
Krause Analytical
Austin, TX USA
2 posts
Page 1 of 1
Who is online
In total there are 42 users online :: 0 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 42 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 42 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science