For your information, the specs of the GC and analysis method are:
GC: GC17-A
Injector: AOC-20i autoinjector
Injection volume: 0.6uL
Injection System: Split injection at 380 degC
Precolumn: Restek MXT Biodiesel TG precolumn, 2m x 0.53mmID x 0.16um
Column: Restek MXT Biodiesel TG column, 15m x 0.32mmID x 10um
Column temperature program: 50-380degC
Column pressure program: 20-65kPa
Detector: FID at 400degC
Normally, baseline is flat with a slight upward drift towards the end of the program as the temperature rises. Yesterday, I firstly noticed the baseline climbing at a considerable rate throughout the course of each sample and the overall analysis. Then ,after about six samples, the baseline passed through a gentle crest mid-analysis, and began to drift downwards. This downward drift continued throughout the remaining six samples. The starting point for the baseline was at the same level as the end point for the previous sample. Despite the drift, there was no noise.
In case of buildup in the column, I tried baking at 410C for two, one hour periods. Column pressure constant at 20kPa. Both times, the chromatogram had a small hump at the start, drifted downward as temp increased, before going through a trough, rising slightly and levelling out at constant temp. No peaks appeared. I retried regular analysis with solvent only, and saw the same trend.
I then cleaned the glass insert of the injector, replaced the septum, silica wool, and the graphite ferrules at the column ends. I don't have replacement ferrules for the injector or connector between the two columns so they are unchanged.
Since then, the baseline has been very erratic and noisy. Ran at 50C for 2h, baseline initially noise free but considerable noise appeared after 1/2-1h. Increased temp gradually to 250C, saw a very sharp, large increase in baseline (up to 13mV, if that is of use). After temp ramping stopped and column steady at 250C, saw a very sharp decrease in baseline, back to about the same level as before ramping, and it continued to drift downward for the next 1/2h that I ran it for. All the while, a very noisy signal.
Do you know what could be the cause/s of this change in baseline behaviour?