I selected GC-TCD using pack Shin carbon ST micropacked column to determine O2 concentration.
To see whether O2 concentration reduced during the time course of shelf life study and correlate to volatile oxidation products. Questions are;

Which O2 standard grade that I need to confirm the O2 peak.
What would be a thing I need to be aware of during this experiment?
In case of syringe, I do have only 50 mL gas tight syringe, can I use a normal 10 uL syringe instead?
Is there any other method else to clarify this phenomenon? or to compare with this method!!
Thanks a lot for all comments! :)
As oxygen is an atmospheric gas, really you should be piping the sample in and injecting via a gas sampling valve. Choose a standard mixture that is close to what you expect to find. I'm not sure what you are trying to measure the oxygen in?
Where can I buy the kit they use in CSI?
If you are trying to measure low levels of O2 (ppm levels) you should not be using a carbon based GC packing.

best wishes,

Rod
3 posts Page 1 of 1

Who is online

In total there are 34 users online :: 1 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry