So, each even-numbered sample matches the other even numbered samples, and each odd numbered sample matches the other odd numbered samples, and this continues indefinitely through a long sequence. Is this still repeat injections from one vial ?
It is very hard to imagine what might cause a regular alternation of peak areas, unless it it something in the programmed sample sequence. What happens if you do six single runs, with a manual start for each one ?
And what are you analysing, and how ?
Peter
Well, this happens indefinitely no matter how many vials you have so if you have five samples, for example, each with 3 injections, each of those samples will have two replicates that match up and one that doesn't. The six injections from one vial was an example to illustrate the point.
The idea of doing six single runs is a good idea. I will try that out in the lab.
The sample is a protein by reverse phase in a gradient method. One other thing is that the retention time of the protein does not change between injections and only one method is used throughout the entire sequence.
One thing that we noticed also is that while the baseline (which increases as ACN percentage increases) is steady through the section where the protein elutes, in a step in the gradient where there is a high ACN wash at the end of the run to clean the column, there is the same alternating pattern. I will post a picture of this in a few minutes when I get back into the lab.