Hi, I am trying to identify some unknown degraded components in a drug with LC/MS. I obtained one very interesting mass spectrum. From mass 100-600, there are five dominant peaks, with mass of ~197, ~295, ~393, ~491 and ~589. The pattern is the distance increase 98Da from the lower mass peak to the higher mass peak. Does anybody know what is the adduct, what may correspond these peaks. Thanks for any suggestion.
emily lee
-
- Posts: 29
- Joined: Fri Nov 12, 2004 10:49 am
Dear Meillid,
Are You using Acetonitrile or any salt in your mobile phase?
Carlos Teixeira
Are You using Acetonitrile or any salt in your mobile phase?
Carlos Teixeira
-
- Posts: 64
- Joined: Thu Feb 24, 2005 5:01 pm
yes. I do use acetonitrile with 0.01%tfa and 0.1%fa
emily lee
-
- Posts: 216
- Joined: Mon Apr 04, 2005 12:26 am
Could it be either phosphoric acid or sulfuric acid?
If so probably eluted in the dead volume of the column.
If sulfuric, would note presence of sulfur isotopes not present if phosphoric acid. If acid would see strong signal in negative ion. At high cone voltages in negative phosphoric fragments to m/z 79and sulfuric to m/ 80.
If so probably eluted in the dead volume of the column.
If sulfuric, would note presence of sulfur isotopes not present if phosphoric acid. If acid would see strong signal in negative ion. At high cone voltages in negative phosphoric fragments to m/z 79and sulfuric to m/ 80.
Sailor
-
- Posts: 29
- Joined: Fri Nov 12, 2004 10:49 am
Good contributions Dear James Little,
But the pattern is the distance increase 98Da from the lower mass peak to the higher mass peak, and what sulfuric species produce this mass?
But the pattern is the distance increase 98Da from the lower mass peak to the higher mass peak, and what sulfuric species produce this mass?
-
- Posts: 64
- Joined: Thu Feb 24, 2005 5:01 pm
in the sample, there is phosphoric acid. but the pattarned spectra is not eluted in the dead volume of the column. it has a RT of about 1:00min and after another unknown peak. please give further advise and thanks
emily lee
-
- Posts: 29
- Joined: Fri Nov 12, 2004 10:49 am
Dear Meillid,
Sorry because I'm late. In my opinion You are having clusters of Fosforic acid plus proton, or (H3PO4)nH+, in your drug analysis. Maybe you change this acid for other?
Have a good day and excellent elutions!!!
Carlos Teixeira
Sorry because I'm late. In my opinion You are having clusters of Fosforic acid plus proton, or (H3PO4)nH+, in your drug analysis. Maybe you change this acid for other?
Have a good day and excellent elutions!!!
Carlos Teixeira
-
- Posts: 216
- Joined: Mon Apr 04, 2005 12:26 am
Did you ever determine the origin of the problem?
Sailor
8 posts
Page 1 of 1
Who is online
In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 32 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science