We are analysing Genotoxic imp. in one of our compund .
( i.e Methyl , Ethyl & isopropyl analouge )
Can anaybody helpus for further developing the method on Headspace , as we have reached approx. 30 ppm response ( LOD valve ) , but we need approx. 5ppm .

Any specific solvent used or column etc
Thanks
Namaste,

On most GCs you should be able to get 5 ppm easily.
Give more details about your methode et instrument.
"If your experiment needs statistics, you ought to have done a better experiment." Rutherford
Hi for GPI analysis you can use split methods and you should be able to go down to 5 ppm. When working with pharma companies we successfully used more polar columns for GPI analysis as most of these analytes do have polar function. Successful phase used are Rtx-200 and the Rxi-17Sil MS,. They have stability up to 350C and show good peak shapes. If you need details plse contact the company.
regards
j de zeeuw
3 posts Page 1 of 1

Who is online

In total there are 36 users online :: 3 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: CharlesPar, Google [Bot], Semrush [Bot] and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry