I am having probelm with 1,1-DCE by EPA Method 8260. All the other compouns including gases look great. The values change from run to run. Any suggestions? Thank you.
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
I have seen this analyte be fouled up by coelution of water and or methanol from the trap.
This is usually a problem with the earlier eluting gases (bromomethane or chloroethane); I have seen it shift on occasion to a later analyte - changing the trap helped in that case.
You can check this by temporarily setting the scan to 15 to 300, as opposed to the more routine 35 to 300 amu. You can verify where the water and methanol are eluting.
This is usually a problem with the earlier eluting gases (bromomethane or chloroethane); I have seen it shift on occasion to a later analyte - changing the trap helped in that case.
You can check this by temporarily setting the scan to 15 to 300, as opposed to the more routine 35 to 300 amu. You can verify where the water and methanol are eluting.
-
- Posts: 81
- Joined: Wed Dec 28, 2011 6:26 pm
Thank you very much tjupille for reply.
I did change the trap, but I still have the same problem. I talked to OI and they suggested to use a freshly made trap! They are shipping me a new trap tomorrow. All the other compound, including brominated look great.
I did change the trap, but I still have the same problem. I talked to OI and they suggested to use a freshly made trap! They are shipping me a new trap tomorrow. All the other compound, including brominated look great.
-
- Posts: 310
- Joined: Thu Jun 18, 2009 7:59 pm
You may have missed the intent of Yama001's response. It's not being caused by the trap - either methanol or water are co-eluting with your 1,1,-DCE. You need to change either your chromatographic conditions or your column.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
Laboratory Director
Krause Analytical
Austin, TX USA
-
- Posts: 81
- Joined: Wed Dec 28, 2011 6:26 pm
Thank you mckrause. You are correct; I did miss the intent and I will look into it.
5 posts
Page 1 of 1
Who is online
In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 21 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 21 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science