I would like to condition 5 columns which are in the same GC (Agilent 6890N) oven, they have the following characteristics:
Col. 1, Col. 2, Col. 5 -> Haysep Q, 3/16" o.d., max. temp. 275ºC (new packed columns)
Col. 3 -> Porapack Q, 1/8" o.d., max. temp. 250ºC (max. temp. assumed according to internet information) (used column that has been stored for some time without end caps or plugs)
Col. 4 -> Molsieve 5A, 1/8" o.d., max. temp. 400ºC (used column that has been stored for some time without end caps or plugs)
According to the information included in the User Manual, I would like to do the following:
Condition the five columns at the same time at a temperature of 250ºC - 30ºC = 220ºC according to the following temperature ramp:
Init temp = 50
Init time = 5
Rate 1 = 15
Final temp = 220
Final time 1 = 720
And establish the following flows:
Col. 1, Col. 2, Col. 5 -> 50 mL/min
Col. 3, Col. 4 -> 25 mL/min
I think that the end of the columns that are normally connected to the detectors (microECD and FID) should be placed out of the oven and I that I should cap the detector fittings. Am I right?
Do you think that the temperature is correct for this type of columns? Time should be longer than 720 min? What do you think about flow rates?
I am doing this because I cannot see any peak with the FID (and methanizer) when trying to detect CH4, CO2 and CO.
I thank you in advance for your help.