OK. We have worked for a couple hours and ran the diagnostics, the detector seems to have failed? Our dynode voltage is -14.93 kV and the multiplier voltage is -6.3 kV. Is this normal. Is there any troubleshooting we can do for the detector to get it working?
Here were the results for the diagnostics:
Begin All Diagnostics Status Test Loop
10:58:41: Multiplier (V) -- Out of range: -6.299
10:58:41: -205 V supply voltage (V) -- Out of range: -189.568
10:58:42: Multiplier (V) -- Out of range: -4.736
10:58:42: -205 V supply voltage (V) -- Out of range: -189.691
10:58:42: Multiplier (V) -- Out of range: -4.736
10:58:42: -205 V supply voltage (V) -- Out of range: -189.568
10:58:42: Multiplier (V) -- Out of range: -6.299
10:58:42: -205 V supply voltage (V) -- Out of range: -189.445
10:58:43: Multiplier (V) -- Out of range: -4.736
10:58:43: -205 V supply voltage (V) -- Out of range: -189.568
10:58:43: Multiplier (V) -- Out of range: -4.736
10:58:43: -205 V supply voltage (V) -- Out of range: -189.568
10:58:44: Multiplier (V) -- Out of range: -4.736
10:58:44: -205 V supply voltage (V) -- Out of range: -189.568
End All Diagnostics Status Test Loop
10:58:44: Multiplier (V) -- Fail
10:58:44: -205 V supply voltage (V) -- Fail
All Diagnostics Status Test Results
+15 V - Pass
+150 V - Pass
+205 V - Pass
+24 V - Pass
+28 V - Pass
+35 V - Pass
+36 V - Pass
+5 V - Pass
-15 V - Pass
-150 V - Pass
-205 V - Fail
-28 V - Pass
8 kV PS voltage - Pass
Ambient temp. - Pass
Analyzer temp. - Pass
Aux gas flow - Pass
Capillary temp. - Pass
Capillary voltage - Pass
Convectron - Pass
Detected RF - Pass
Dynode voltage - Pass
Entrance lens - Pass
Intermultipole lens - Pass
Ion gauge - Pass
Main RF DAC - Pass
Multiplier setting - Pass
Multiplier voltage - Fail
Multipole 1 offset - Pass
Multipole 2 offset - Pass
Multipole RF mod. - Pass
Multipole det. RF - Pass
RF amp. output - Pass
RF det. temp. - Pass
RF gen. temp. - Pass
RF modulation - Pass
Sheath gas flow - Pass
Trap DC Offset - Pass
Tube/gate lens - Pass
Starting All Diagnostics Scan Tests
10:58:44: Start scan readback test on device Auxiliary amplitude (V)
-- 0 to 83.2
10:58:47: Scan readback test ended
10:58:48: Result: PASSED
10:58:48: Start scan readback test on device Main RF DAC (16-bit) -- 0 to 65535
10:58:52: Scan readback test ended
10:58:52: Result: PASSED
10:58:52: Start scan readback test on device Vernier det. RF amp. (V)
-- 0 to 65535
10:58:56: Scan readback test ended
10:58:56: Result: PASSED
10:58:56: Start scan readback test on device Vernier RF DAC (16-bit)
-- 0 to 65535
10:59:00: Scan readback test ended
10:59:01: Result: PASSED
10:59:01: Start scan readback test on device Multipole RF DAC (V) -- 0 to 1000
10:59:04: Scan readback test ended
10:59:05: Result: PASSED
10:59:05: Start scan readback test on device Multipole 1 offset (V)
-- -132 to 132
10:59:09: Scan readback test ended
10:59:09: Result: PASSED
10:59:09: Start scan readback test on device Multipole 2 offset (V)
-- -132 to 132
10:59:13: Scan readback test ended
10:59:13: Result: PASSED
10:59:13: Start scan readback test on device Multipole lens (V) -- -132 to 132
10:59:17: Scan readback test ended
10:59:17: Result: PASSED
10:59:17: Start scan readback test on device Multipole det. RF amp.
(Vp-p) -- 0 to 1000
10:59:21: Scan readback test ended
10:59:22: Result: PASSED
10:59:22: Start scan readback test on device Trap Offset (V) --
-133.786 to 133.882
10:59:25: Scan readback test ended
10:59:26: Result: PASSED
10:59:26: Start scan readback test on device Tube gate (V) -- -200 to 198.01
10:59:30: Scan readback test ended
10:59:30: Result: FAILED
10:59:30: Start scan readback test on device Multiplier (V) -- 0 to -2200
10:59:43: Scan readback test ended
10:59:43: Result: FAILED
10:59:43: Start scan readback test on device Aux/sweep gas flow (arb)
-- 0 to 60
11:01:17: Scan readback test ended
11:01:18: Result: PASSED
11:01:18: Start scan readback test on device Sheath gas flow (arb) -- 20 to 100
11:02:53: Scan readback test ended
11:02:54: Result: PASSED
11:02:54: Start scan readback test on device Capillary Voltage (V) --
-132 to 132
11:02:57: Scan readback test ended
11:02:58: Result: PASSED
11:02:58: Final result: FAILED
Thanks!!!!!