AICMM » Fri Jan 06, 2012 4:13 pm
vestelshirley,
A shincarbon might do most of this job, depending. I set one up for a very similar application, pyrolysis of sugar chaff. But it is not quite that simple.... At reasonable temperatures, it will not be possible to separate O2 and argon (which may not matter that much and which is probably better done with an oxygen meter anyway) and it will take a long time (and a temp program) to get the butane off the Shincarbon. The better way to do this is to run two channels, one for the fixed gases and the other for the hydrocarbons (perhaps a Shincarbon and a Q) in parallel. You will still inject heavies on the Shincarbon though, so ideally this is set up as a backflush to vent configuration.
The other problem, as Rodney noted, is that the hydrogen and the other fixed gases will probably require two TCD's or a single HID to analyze. Getting to be a bit hairer of an instrument, but still very doable.
Best regards,
AICMM