Hi,

I'm quite new to this and my 5972 MS is not functioning the way I'd like it to function. I don't know what you need to know to make a verdict, but there are some things I need to point out.

autotune is failing. Even other tune files won't finish, saying the peaks are too wide and the offset is at maximum... at other times it will say it can't get a stable peak width or that there isn't enough signal to start. And when I run it at 2500EV it gives me this....

Image

We're at our wits end...
Please help
Have you tried running at a lower voltage? 2500 seems pretty high to me.

Usually failing tune is a sign you need to do some source cleaning and maintenance and possibly switch/replace a filament.
Nathan Valentine
Purge and Trap Product Line Manager
Teledyne Tekmar
http://www.teledynetekmar.com
http://www.teledynetekmarblog.com/
We just cleaned the source as mentioned in the manual and we even had an agilent specialist look at it for us.

We've just replaced the filaments too...
And has anyone got any clue why the baseline drops after 17 minutes, because it does so in every run?
o0kim0o,

1) Replace your EM. Then see what the tune does.

2) Replace your repeller and ion focus lens.

Hopefully this will get your tune to finish. When your autotune starts, how much 69 do you see?

3) build a real method, not the default. Set up real/realistic flows and split/splitless times. Keep it isothermal for now (say 50 C) but with a real column flow at less than a ml/min.

Then we go from there....

Best regards,

AICMM
You either have a lens shorting out or you have a quad driver issue. Clean your source, and make doubly certain that you get it assembled correctly! I don't think you have an EM issue - peak width issues aren't resolved by increasing EM voltage. You may also have an EM issue, but that's not causing the peak width error.

You may have an issue with the quads on the '72. The glass quads are garbage. If you're good with hardware and have MS experience I'd recommend solvent rinsing them - if they get oil on them you can have all kinds of weird tune experiences. I've also seen them get so contaminated that you can't bring them back. It normally manifests itself as a mass issue - you can't set 219, but you have boatloads of 213. You can see this in scan - open the PFTBA valve and run a scan to see what your spectrum looks like.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
Well...
Now we can't even get the MS to turn on.
It keeps giving fault 8 even though we checked the filaments and the cables.
Everything seems to be okay, so we have no idea what's wrong :(
7 posts Page 1 of 1

Who is online

In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry