I am analysing 5-aminosalicylic acid (5-ASA) on a mixed-mode column using TFA/water/acetonitrile as mobile phase. The sample is dissolved in an EDTA solution in water (pH about 7). The reason for using EDTA is that 5-ASA is very easily oxidised, and EDTA makes the solutions stable for weeks.
The problem is that when I start a new sequence, I get a quite a lot of area drift (increasing areas). It can take up to 10-15 standard injections before the areas become stable. Another strange thing is that the response factor of the standard is quite different in different sequences (using the same instrument and mobile phases). It can shift plus/minus 5% from one day to another.
I have performed quite a lot of trouble shooting already:
- I replaced the column with some PEEK tubing = same area drift observed
- I have used a metal-free LC-system = same area drift observed
- I have tested to siliconise the needle and the LC-vials. It seemed to increase the response factor, but it didn't remove the drift.
- The area drift seems to get worse if I inject the standard from the same vial, compared to performing one injection/vial. The drift is however too large and fast to be explained by evaporation.
All ideas are welcome!!