The key is in CPG's last question (injection of water or methanol) with one addition: what happens if you inject mobile phase.
What all of this is leading up to is that what you are describing sounds very much like a "syatem peak". More details about your mobile phase and detection might help to pin it down further. Another (more remote) possibility is an artifact from the "refernce wavelength" function of a DAD, but the fact that it shows up in a blank makes it less likely.
In any case, if it doesn't interfere with a peak of interest, you can always deal with it the old-fashioned way: ignore it.
I had the same problem as the OP. The negative dip can be seen towards the end of the chromatogram, and I have not changed my methods. This phenomenon is not column related (seen it on a new column).
My answer to CPG's questions are: sometimes Yes (the HPLC is used for several methods), Yes, and Yes (methanol)

. Haven't tried Tom's additional question though.
Flushing the system with MeOH seems to have no effect, and this effect is more prominent at lower wavelength (225nm) than higher wavelengths (I am using a PDA/DAD detector). The reference wavelength or Blank subtraction is not enabled.
My mobile phase is comprised of 0.1% TFA in ACN and 0.1% TFA in H2O. The instrument is a Waters Alliance 2695 with 2998 PDA detector.