Two Phosopholipds, same retention time...

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Hi

I have two phospholipids that are very similar in MW (one is ~1350, the other is ~1495) that I am trying to separate via UPLC. I have tried several different things so far in terms of gradient and buffer, but they both have the same exact RT no matter what I try.

Any suggestions?
Hi,
Could you tell us more about those phospholipids? Is the only difference a molecular weight? And it would be because difference in length of lipid chain I suppose?

Some time ago I saw some HPLC separations of closely-related lipophilic molecules with even smaller (or none) difference in MW using C30 column - have you tried that?
Hi,
Could you tell us more about those phospholipids? Is the only difference a molecular weight? And it would be because difference in length of lipid chain I suppose?

Some time ago I saw some HPLC separations of closely-related lipophilic molecules with even smaller (or none) difference in MW using C30 column - have you tried that?
Hi

Lipid 1: Phospatidylinositol3,4,5-triphosphate MW= 1795
Lipid 2: Phospatidylinositol4,5-bisphosphate MW= 1513

The column I am using now is: Phenomenex Kinetex C8 2.1x100mm 2.6um, I have not yet tried a C30 column.
Other idea - they have different charge so perhaps you could try ion exchange? With organic solvent in mobile phase to decrease interactions of lipid chains?
What pH and buffer are you using? I would guess pH 7-8, would give you the best chance for selectivity. Also, what organic modifiers have you tried?
A. Carl Sanchez
several things that can help
use a polar embedded column like the HP c-18 from sepax
http://www.sepax-tech.com/application_notes/OD1001.pdf

or you could try to reduce the temp, in the oven,
10 or even 5 degrees make better for separation, i don't think the column will be capable of handling the pressure on UHPLC

you could also try a cholester from nacalai
the C-30 also works best at lower temp, of 20 or less

you could also try a HILIC column i know it works also for the nucleotide mixture
6 posts Page 1 of 1

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry